Re: Несостятельность фиксации
послал Alexander Chelnokov 07 Август 2003, 12:54
|
> 1. Whould be interesting to see pre-op X ray.
Attached.
> 2. The reduction is rather - mal-reduction: Syndesmosis screw is
> pushing against the tibia and widening the Tibio-fibular space and
In short - no, it was pure position screw, fragments didn't move after its insertion.
> 3. Medial distal tibial fracture is fixed with a plate which is know to
> cause skin necrosis - it is a big problem.
In this case the wound healed uneventfully.
> For all of the above: do CT scan( can be done even in the presence of
> the metal), and remove the hardware( plates and screws). Reduce the
Something similar we discussed after first glance but recognized the plan as too risky, this is why the case is being discussed... Anyway
thanks for your input. Looking forward to hearing from you.
|